Dear Eydís,
With a low throughput instrument one must be pragmatic. Best is to use “All-genes” or “All samples” plate designs, because then you don’t need interplate calibration. In your case this is analyzing all samples for GOI_1 on one plate, all samples for GOI_2 on a second plate, RG_1 on a third etc. NTC’s can be run on a separate plate. RG- should be included with the positive samples, but it’s not critical, because the inter-plate calibration is in general only minor, and you only use those samples to have an indication of LOD (when limited by the RT performance).
If you cannot use an “All-samples” design or want to keep flexibility you can include IPC. These should be uncomplicated samples, which robust qPCR assay that give rather low Cq with minimum variation (low SD). You should preferably run the IPC in triplicates. You need one IPC for each channel of the instrument (not one for each assay).
Have you considered outsourcing the expression profiling? TATAA Biocenter (
www.tataa.com) , for example, runs customer samples on 384-well plates or even microfluidic cards that take 10000 parallel samples. That would save you time and probably also money (reagent costs are lower on high throughput instruments).
Good luck!