It is not a good idea to perform separate standard curves on each plate/day. If you do that you will see a variation in efficiency, but this will not reflect true day-to-day variation, but rather depend on the particular noise you experience every time; unless, of course, you run very extensive standard curves (28 or standard points) each time. This is necessary if you introduce a change that can influence performance, such as new batch of primers, new master mix, new technician etc. If you don't expect systematic difference it is better to perform one single very accurate standard curve and use it to standardize all experiments. But you must run an inter-plate calibrator (IPC) to remove run-to-run variation. It is also advisable to run a couple of controls each time (there are recommended concentrations for those). I suggest using GenEx for the standard curve analysis, since it will also estimate the accuracy of the standard curve, identify any outliers, determine the dynamic range and, if you need, also LOD and LOQ. You can make your own IPC or you can get it from eg. TATAA Biocenter (
www.tataa.com). You find also info on GenEx, which is from MultiD, on TATAAs website.
Good luck